是德科技
李桂香
8年用户 171经验值
改进高功率放大器设置的校准
* PNA型号* E8363C;
A.09.42.22;
选项010,014,P02 *目标:*表征100W放大器的增益,绝对输出功率,VSWR,增益压缩。
*测试设置:*类似于应用笔记1408-10的图4,除了我不需要前置放大器和我的输出测试耦合器是30 dB。
我也没有在rcvr B in和Cplr ARM之间使用10 dB。
所需的端口1功率(DUT输入)将从-25dBm到0dBm。
* PNA配置:* 101分,IFBW = 1kHz,Avg = 7 *问题1:*校准后,S21代替DUT不平坦(> 1.5 dB)我首先使用功率计/传感器在端口上进行功率校准
1(-25dBm至0 dBm,步长为1 dB)。
然后我用eCal模块(N4691-60004)更换我的DUT并进行全2端口S-param校准(端口功率为-10 dBm)。
接下来我把_thru放入我的DUT_然后看S21(dB)并且没有得到我期望的平坦度。
我尽量减少测试电缆的移动,并小心所有连接。
如果我做了响应直通,那么我可以获得平坦度我有不同的IFBW,平均值,初始端口功率水平,测试电缆,测试耦合器,具有类似的结果。
*问题2:*我可以显示DUT的绝对功率(抵消测试耦合器和电缆的输出损耗)吗?
为了查看DUT的绝对输出功率,我正在测量接收器B / 1.0(源= P1)。
我可以偏移此数据(通过频率)来显示我的DUT的电源输出吗?
*问题3:*您能否确认我的PNA配置不允许内部调整接收器B的衰减?
另外,我是否需要在外部添加衰减(cplr arm到RCVR B in)以减少Receiver B压缩?
任何帮助表示赞赏。
以上来自于谷歌翻译
     以下为原文
  *PNA Model* E8363C; A.09.42.22; Options 010, 014, P02
*Goal:* characterize Gain, Abs Output Power, VSWR, Gain Compression of a 100W amplifier.
*Test setup:* similar to Fig 4 of App Note 1408-10 except I do not need the preamp and my output test coupler is 30 dB.  I also am not using the 10 dB between the rcvr B in and Cplr Arm.
Desired Port 1 power (DUT input) will be from -25dBm to 0 dBm.
*PNA config:* 101 pts, IFBW = 1kHz, Avg = 7
*Issue1:* S21 with thru in place of DUT is not flat (>1.5 dB) after calibration
I first conduct a power calibration with my power meter/sensor on Port 1 (-25dBm to 0 dBm in 1 dB steps).
Then I replace my DUT with the eCal module (N4691-60004) and do a full 2 port S-param calibration (at port power of -10 dBm).
Next I put a _thru in for my DUT_ and look at S21 (dB) and do not get the flatness I would expect.  I try to minimize test cable movement and am careful with all connections.
If I do a Response Thru cal then I can get flatness < 0.05dB, but thought full 2 port would be the best cal and also want to meas other S-params.
I have varied IFBW, Averaging, Initial port power level, test cables, test coupler, with similar results.
*Issue2:* Can I display the absolute power at the DUT (offset the output losses of the test coupler and cables)?
In order to see the absolute output power of the DUT, I'm measuring Receiver B/1.0 (source = P1).  Can I offset this data (over freq) to display the power out at my DUT instead?
*Issue 3:* Can you confirm that my PNA configuration does not allow for internal adjustment of the Receiver B attenuation?   Also, will I need to add attenuation externally (cplr arm to RCVR B in) in order to reduce the Receiver B compression?
Any help is appreciated.
附件
回帖(4)
刘华湘
2018-10-10 17:33:25
> {quote:title = GT_Test写道:} {quote}> *问题1:* S21代替DUT在校准后不平坦(> 1.5 dB)>我首先在端口上使用功率计/传感器进行功率校准
>如果我做了响应,那么我可以得到平坦度我有不同的IFBW,平均值,初始端口功率电平,测试电缆,测试耦合器,具有类似的结果。
>因为您将30 dB耦合系数外部耦合器的耦合臂连接到端口2上的测试端口,您基本上在端口2路径中添加了30 dB衰减器。
当您使用-10 dBm端口电源执行完整的2端口S参数ecal时,连接到端口2的ecal端口的功率电平为-40 dBm,并且在校准期间接收器B测量的任何反射信号将为
-70 dBm或更低。
如果你想在这种配置中获得更好的2端口校准,你必须解除端口功率并将端口2的功率设置为尽可能高而不会不平衡并减少IFBW(和/或增加平均值
在cal)期间。
您的另一个选择是只执行EnhancedResponse cal,这将允许您测量S11和S21,但不能测量S22。
这是大多数具有高功率设置的人通常所做的事情,因为您将测量的S22基本上是30 dB耦合臂的匹配,而不是放大器输出端口的匹配。
但是,如果您真的想要一个良好的完整2端口校准,那么您将需要设置一个双外部反射计。
下面是PNA-X上一个这样的设置的示例框图,但这个概念也适用于您的E8363C PNA。
在下面的示例中(来自高功率应用笔记1408-19),您可以忽略输入外部放大器,但如果您想要进行完整的2端口校正,则输出双反射计设置就是您所需要的。
!HTTPS://dl.dropboxusercontent.com/u/87949221/Forum/Thread_119233/HighPowerTestset.png!
以上来自于谷歌翻译
     以下为原文
  > {quote:title=GT_Test wrote:}{quote}
> If I do a Response Thru cal then I can get flatness I have varied IFBW, Averaging, Initial port power level, test cables, test coupler, with similar results.
because you are connecting the coupled arm of your 30 dB coupling factor external coupler to the test port on port 2, you are basically adding a 30 dB attenuator in the port 2 path.  when you do your full 2-port S-Parameter ecal with -10 dBm port power, the ecal port that is connected to port 2 is seeing a power level of -40 dBm and any reflected signal measured by receiver B during the cal will be -70 dBm or lower.  if you want to get a better 2-port cal in this configuration, you'll have to uncouple the port powers and set the port 2 power as high as it can go without going unleveled and reducing the IFBW (and/or increasing the averaging during the cal).  Your other option is to just perform an EnhancedResponse cal, which will allow you to measure S11 and S21, but not S22.  this is what most people with high power setups usually do, since the S22 that you will measure will be essentially the match of the 30 dB coupled arm and not the match of your amplifier's output port.
However, if you really want a good full 2-port calibration, then you will need to setup a dual external reflectometer.  below is an example block diagram of one such setup on a PNA-X, but the concept applies to your E8363C PNA as well.
in the example below (from the high power app note 1408-19) you can ignore the input external amplifier, but the output dual reflectometer setup is what you need if you want to do a full 2-port correction.
!https://dl.dropboxusercontent.com/u/87949221/Forum/Thread_119233/HighPowerTestset.png!
李桂香
2018-10-10 17:41:57
非常感谢帮忙。
增强的响应校准给了我很好的结果。
我的幅度变化是
以上来自于谷歌翻译
     以下为原文
  Thanks so much for the help.  The enhanced response calibration gave me good results.   My amplitude variation was <0.2dB after this calibration and I'm able to meas S11 as well.
马沃源
2018-10-10 17:51:08
你好Daras等人,
我也对使用RFOW探针进行高功率校准感兴趣。
目标#1:表征放大器的增益和大信号S参数@ + 21dBm。
DUT的输入电平:+ 11dBm。
目标#2:表征增益,绝对输出功率,VSWR,增益压缩
DUT的输入电平:-10dBm至+ 11dBm
我有一个测试装置,包括一个驱动器,DUT输入和输出端的双向耦合器和衰减器(20dB),以限制分析仪接收器输入的功率远低于-20dBm。
我使用应用笔记1408-10进行测试设置,但不清楚校准需要哪些步骤。
有人可以帮助我吗?
我认为对于目标#2,我需要做以下事情:
你能把这份文件寄给我吗?
谢谢。
以上来自于谷歌翻译
     以下为原文
  Hello Daras and et al,
I am also interested in doing high power calibration but using RFOW probes.
Objective#1: Characterize Gain and large signal S-parameters of an amplifier @ +21dBm.
Input level to DUT: +11dBm.
Objective#2: characterize Gain, Abs Output Power, VSWR, Gain Compression
Input level to DUT: -10dBm to +11dBm
I have a test setup which includes a driver, bi-directional couplers at the input and output of the DUT and attenuators (20dB) to limit power to the analyser's receiver inputs well below -20dBm.
I used the App note 1408-10 for the test setup, but it's not clear what steps are required for calibration.
Can someone please assiste me?
I take it that for Objective#2, I need to do the following:
Can you please send me this document?
Thank you.
马沃源
2018-10-10 18:07:55
引用: xj5548 发表于 2018-10-1 09:51
你好Daras等人,
接收器电平应低于-20dBm,以尽量减少失真。
但另一方面,太低的信号会影响校准精度。
接收器的动态范围或最小可检测水平是多少?
以上来自于谷歌翻译
     以下为原文
  The receiver levels should be below -20dBm to minimise distortion.  But on the other hand too low signal affects calibration accuracy.
What is the dynamic range or minimum detectable level is for the receivers?